So why do I get this white banding effect whenever i try to scan an image?
Is my scanner broken?
Image Quality
Moderator: LSI_Ketelhohn
- LSI_Ketelhohn
- LSI Staff
- Posts: 4283
- Joined: Wed May 28, 2008 11:19 am
- Scanner: all
- Location: Kiel, Germany
- Contact:
Re: Image Quality
Dear customer,
looks strange as if something reflects the light from the transparency unit.
Did you check for any foreign matter in the scanner?
Maybe something got stuck to the scanner sensor.
kind regards,
Arne Ketelhohn.
looks strange as if something reflects the light from the transparency unit.
Did you check for any foreign matter in the scanner?
Maybe something got stuck to the scanner sensor.
kind regards,
Arne Ketelhohn.
Re: Image Quality
Well I cleaned everything again but I still get the same result.
Can I clean this? If so, how?
I actually used my V700 once before.
I scanned 1200x35mm without any problems.
But now that I need it again I get this effect....
I use SilverFast (Epson-SE).8ba (vers. 6.6. 1r2b)
with the following settings:
I also can't seem to select a "professional" mode.
I only have basic & standard. (why?)
LSI_Ketelhohn wrote:Maybe something got stuck to the scanner sensor.
Can I clean this? If so, how?
I actually used my V700 once before.
I scanned 1200x35mm without any problems.
But now that I need it again I get this effect....
I use SilverFast (Epson-SE).8ba (vers. 6.6. 1r2b)
with the following settings:
I also can't seem to select a "professional" mode.
I only have basic & standard. (why?)
- LSI_Ketelhohn
- LSI Staff
- Posts: 4283
- Joined: Wed May 28, 2008 11:19 am
- Scanner: all
- Location: Kiel, Germany
- Contact:
Re: Image Quality
You would have to contact the manufacturer on how to clean the device.
Your settings look OK but you are using an outdated version.
There is a more current update available from our download website.
SilverFast SE only offers basic and standard mode.
The full professional solution is available in SilverFast Ai Studio.
kind regrads,
Arne Ketelhohn.
Your settings look OK but you are using an outdated version.
There is a more current update available from our download website.
SilverFast SE only offers basic and standard mode.
The full professional solution is available in SilverFast Ai Studio.
kind regrads,
Arne Ketelhohn.
Re: Image Quality
I've tried it again with the current version SE version. 6.6.2r5 but with the same results.
I don't seem to have gotten an Ai version with my Epson V700.
i'm a bit skeptical on buying the Ai version since I don't know whether my problem will be solved or not.
I don't seem to have gotten an Ai version with my Epson V700.
i'm a bit skeptical on buying the Ai version since I don't know whether my problem will be solved or not.
- LSI_Ketelhohn
- LSI Staff
- Posts: 4283
- Joined: Wed May 28, 2008 11:19 am
- Scanner: all
- Location: Kiel, Germany
- Contact:
Re: Image Quality
A couple things to locate the issue:
Use epson scan to see if the error repeats.
Do you see the same issue on reflective scans ?
Is the band always in the exact same location (at the pixel level) ?
It appears the gain is elevated for a few bands of pixels - Check the calibration area (under the case holding the platen glass where the scan head is parked to see if there is any debris).
Frank
Use epson scan to see if the error repeats.
Do you see the same issue on reflective scans ?
Is the band always in the exact same location (at the pixel level) ?
It appears the gain is elevated for a few bands of pixels - Check the calibration area (under the case holding the platen glass where the scan head is parked to see if there is any debris).
Frank
Re: Image Quality
degrub wrote:A couple things to locate the issue:
Use epson scan to see if the error repeats.
Do you see the same issue on reflective scans ?
>>> This is a screenshot of a pre-scan so its bad BUT at least no thick white line.
Is the band always in the exact same location (at the pixel level) ?
>>> i seems to be in the exact same place yes.
It appears the gain is elevated for a few bands of pixels - Check the calibration area (under the case holding the platen glass where the scan head is parked to see if there is any debris).
>>> It looks clean. And the reason why I responded so late is because my scanner was being 'repaired' at Epson's. Well they tested every form of scanning (dia, film, reflective) and found NOTHING...
So they did just send it back and they preventively cleaned all plates and cleaned the scanner from the inside out.
So at least It should be clean.
Frank
So reflective scans work.
Could it be the cable between the top unit and the base unit?
Thanks
kenny
- LSI_Ketelhohn
- LSI Staff
- Posts: 4283
- Joined: Wed May 28, 2008 11:19 am
- Scanner: all
- Location: Kiel, Germany
- Contact:
Re: Image Quality
It could be the cable.
But it does not seem like a common issue.
Normally a defective connection causes the scanner to loose contact to the transparency unit.
This is unusal bvehavior.
The white stripes look as if far too much light gets projected at the image.
You can even see the film structure through the holder.
Is taht a standard holder or was it modified?
kind regards,
Arne Ketelhohn.
But it does not seem like a common issue.
Normally a defective connection causes the scanner to loose contact to the transparency unit.
This is unusal bvehavior.
The white stripes look as if far too much light gets projected at the image.
You can even see the film structure through the holder.
Is taht a standard holder or was it modified?
kind regards,
Arne Ketelhohn.