IT8 Calibration Problem

General topics about imaging

MBane
Visitor
Posts: 3
Joined: Wed Nov 25, 2015 6:28 pm
Scanner: Braun FS-120
SilverFast Product: Ai Studio

IT8 Calibration Problem

Postby MBane » Fri Nov 27, 2015 9:18 pm

I am trying to (re)calibrate my Braun FS-120 Scanner. I have a Fuji 6x7 IT8 target. I have acquired the relevant reference file for the target. However, I am going around in circles. Whatever I do I end up with the same error as shown in the attached image. What is the problem? Thanks
Attachments
probs.jpg
probs.jpg (185.66 KiB) Viewed 643 times

User avatar
LSI_Ketelhohn
LSI Staff
LSI Staff
Posts: 4283
Joined: Wed May 28, 2008 11:19 am
Scanner: all
Location: Kiel, Germany
Contact:

Re: IT8 Calibration Problem

Postby LSI_Ketelhohn » Tue Dec 01, 2015 11:43 am

Dear customer,

You can grab the corners of the red frame to move them to the correct position.
The black marker must be positioned on the black field on the grey scale and the white marker on the white field of the greyscale.

Kind regards,
Arne Ketelhohn.