"iT8 Calibration" calibration or profile creation?

Postby matt » Wed Jun 06, 2007 11:48 pm

Hi

I am using Silverfast AI Studio with a LinoType Hell Tango Scanner... i was just wondering if the iT8 Calibration tool is doing anything aside from creating a custom icc profile for my scanner and loading it into the CM settings?

is it safe to assume that the "iT8 calibration" does not modify the hardware in any way or change the way my scanner functions? instead it just creates a scanner profile?

Thank you...

User avatar
LSI_Noack
LSI Staff
LSI Staff
Posts: 1432
Joined: Wed Oct 22, 2003 1:30 pm
Location: Kiel, Germany
Contact:

Postby LSI_Noack » Tue Jun 12, 2007 10:53 am

Dear Matt

indeed that's what the IT8 calibration does:
(a) reading color data of target from scanner
(b) comparing to values found in reference file
(c) generating ICC profile for your scanner
(d) saving profile to the relevant system directories
(e) activating scanner profile in SilverFast's CMS settings

That's it, nothing more - hardware isn't set up / changed in any way.
It's save.

Best regards
Sonny Noack
- Manager Technical Support, LaserSoft Imaging AG -